BIR: determining if a slice of a sagittal image is to the left or right of the medial
I'm implementing the requirement for BIR that says I should indicate
whether the current sagittal image is on the left or right hand side
of the medial sagittal line.
It's easy enough to determine if an MRI image is sagittal from the
directional cosines. Is this simply a matter of checking the Image
Position Patient attribute and if it is > 0 the image is to the left
and if it is < 0 it is to the right?
I'm assuming for simplicity that the patient positions is HFS.
Best regards,
Bill Prada
|
3/30/2010 10:06:13 PM
|
0
|
"b...@sorna.com" <b...@sorna.com>
|
|
|
CT Image Tags - Thickness vs. Location
Hi all,
I am looking at a patient from the OsiriX samples (http://
pubimage.hcuge.ch:8080/) called CEREBRIX. The CT series has 174 slices
and each slice has a Slice Thickness of 2, but the Slice Location
values differ by only 1 between sequential slices (I'm assuming the
thickness goes 1 mm in each direction from the location). For example:
Instance 1: Slice Location = 30.5; Slice Thickness = 2
Instance 2: Slice Location = 31.5; Slice Thickness = 2
Instance 3: Slice Location = 32.5; Slice Thickness = 2
My question is if DICOM gives explicit instructions, that I've missed
in the
|
3/30/2010 2:31:20 PM
|
0
|
Chad <cnel...@prismclinical.com>
|
radiation dose exploitation: MPPS or SR
Hi all,
I'm willing to make an extended use of the Radiation Dose provided by
different modalities along the Hospital.
Actually, this information is not widely used or not as much as it
could be, and due to its potential interest I've decided to study the
way how to exploit it.
I've few knowledge about radiation dose information and its exchange
among PACS environment but all I have seen so far by now is that the
current way for communicating this type of info (MPPS messages) is no
longer "recommended" anymore but instead to use proper SR objects (at
least on XRay and CT scenario
|
3/29/2010 4:12:11 PM
|
1
|
Jordi <jhugu...@gmail.com>
|
Worklist
Must a Worklist-Server only send the items referred to the modality?
Our Worklistserver sends the whole List to all connected Modalities. In
some cases we have double Study-UIDs in different Modalities!!
The Vendor says, that this is no problem of the Worklistserver. They
can't setup in their software, that only the items for the modality (ex.
CT or MR) are transferred.
Some meanings about this topic?
|
3/27/2010 9:34:23 AM
|
0
|
Markus Schneider <markus_schneide...@gmx.de>
|
Query modality (Scanner) from other DICOM nodes
Hi All,
I would like to know your view on Querying modality from other DICOM
nodes. Is this really needed? Should the modality (scanner) support
DICOM Q/R?
My impression is that, this can lead to a incomplete/inconsistant data
in a hospital workflow. For example, if a system retrieves study from
modality while the scan is half way, then it provides only an
incomplete data.
Is there any workflow in which the query/retrieve to a modality needs
to be supported in real hospital workflow.
Thank you,
Sujesh
|
3/26/2010 3:47:33 AM
|
0
|
Sujesh <sujuparava...@gmail.com>
|
EXPORT SELECTION TO DICOM FILE
Hi to All:
I would want to ask you if you know a software that permits to realize
a selection in a DICOM image (CT image) and export this selection in a
new DICOM image??
For example, I have DICOM image 420mm X 420 mm size per frame but I
need only a 200mm X 200mm region per frame.
Best Regards
Oscar Garc=EDa
|
3/26/2010 2:54:29 AM
|
0
|
Oscar Garcia <ojgarc...@gmail.com>
|
help with GE MRI image voxel positions
Dear All,
The question is regards GE MRI image.
I have two MRI images of the same slice. They were acquired with
different FOV: one 230mm X 230 mm, and the other 160mmX 160mm. But
they are not co-centered.
I know the ImagePositionPatient and ImageOrientationPatient for the
first MRI image from the DICOM header, and the Center (cx, cy, cz in
mm) of the 2nd image from the P-file (GE's own format).
I want to find out the position of the Center of the 2nd image on the
1st image.
The algorithm is simple:
1. get the IPP2C vector -- a vector from ImagePositionPatient to
Center
|
3/25/2010 8:06:54 PM
|
0
|
Xiehe Cheng <xiehec...@gmail.com>
|
Is there a way to go to a particular topic?
I just recently started to read this group. Currently there are
8000+ topics. Do I have to read them sequentially back to
the beginning or there is a better way to jump to an arbitrary
topic number?
|
3/25/2010 2:25:08 PM
|
0
|
Donald Li <lishuch...@gmail.com>
|
digital signature and watermark
Hi all, I'm sorry for my bad English but I need to know one thing. In
this period I'm developping a software that could sign a DICOM image
with a digital signature and then could verify the integrity and the
authentication of the image, The standard suggests to hide the digital
signature in the tags of the DICOM file, but for my goals this method
is not enough secure. So I decide to test the technique of the
watermark that seems to be the best for my goals. My question is:
talks the standard about this in some part? If yes, defines the
standard a limit of the minimum PSNR (peak signal
|
3/24/2010 8:36:50 AM
|
0
|
signo88 <mirko.signor...@nsbusiness.it>
|
Which Relationship Type (HAS Properties or HAS CONCEPT MOD) to use for US Vascular measurement for CID 224?
Hi Sir/Madam:
For US Vascular Measurement (e.g NeckArt_LT_CCA_Prox_PeakSysVelocity)
you can specify the selection method in CID 224 e.g. code: 121411 -
Most recent value chosen, for the last measurement.
What I want to know is which relationship type (HAS Properties or HAS
Concept Mod) relationship type can be used to describe that this
Peaksystolic Velocity measurement is the "Most recent value
chosen"?
I see in the past, for OB measurements, the "Mean" of a measuremet e.g
Head Circumference Mean has always been sent using the "HAS Concept
Mod" relationship type.
e.g Head
|
3/23/2010 11:57:16 PM
|
0
|
"enyaw_2...@hotmail.com" <enyaw_2...@hotmail.com>
|
DX: ViewPosition - how to obtain?
Greetings,
I am looking at some DICOM data for DX images. I have a single series
listed as PA, which contains 3 images.
In one image:
SeriesDescription is PA
But, it is obvioulsy an Oblique view, and both SeriesDescription and
ProtocolDescription DICOM data confirm this with the value of OBL.
However, ViewPosition (0018,5101) is not present, so how can I
accurately determine if the ViewPosition is RLO or LLO from the DICOM
data alone?
For this example
(0020,0020)PatientOrientation is L\F
(0020,0062) ImageLaterality is U
but,
(0020,0060) Laterality is also not presen
|
3/22/2010 2:50:08 PM
|
0
|
hangingProtocol Research <hangingprotocolresea...@gmail.com>
|
Java Dicom Viewer Lite for DCM4CHEE PACS
It seems the Turyon Projects guys have released a new version of their
Java Dicom Viewer Lite for DCM4CHEE PACS and can send images using
WADO to OsiriX nodes. It's free/libre software under GPL v3 license.
Images can be downloaded using WADO over the Internet and once
received can be delivered automatically to OsiriX, so there is no need
to mess around with PACS authorized node list, DICOM protocol, router
ports, VPN and so on.
Take a look at http://www.turyon.com
I think there are people here interested in sharing images and reports
(SR) over the Internet in such an easy way. I had
|
3/22/2010 9:45:12 AM
|
0
|
irimi1977 <irimi1...@gmail.com>
|
Type of DICOM Attributes
Hi All,
I wanted to know the DICOM tag TYPES for the following tags,
Please help me
[0x00080092] =3D eST; //Referring Physician's Address
[0x00080094] =3D eSH; //Referring Physician's Telephone
number
[0x00100050] =3D eSQ; //Patient's Insurance Plan Code
Sequence
[0x00101005] =3D ePN; //Patient's Birth Name
[0x00101040] =3D eLO; //Patient's Address
[0x00101060] =3D ePN; //Patient's Mother's Birth Name
[0x00101090] =3D eLO; //Medical Record Locator
[0
|
3/22/2010 8:51:05 AM
|
0
|
1997KarnatakaGroup <bagiha...@gmail.com>
|
New SOP classes
Hi,
Which vendors / applications / modalities are generating the following
SOP classes:
Breast Tomosynthesis Image Storage SOP Class
1.2.840.10008.5.1.4.1.1.13.1.3
Basic Structured Display Storage
1.2.840.10008.5.1.4.1.1.131
Enhanced MR Color Image Storage SOP Class
1.2.840.10008.5.1.4.1.1.4.3
Or can someone provide sample data for these Storage classes?
Thanks,
SJ
|
3/22/2010 6:26:12 AM
|
0
|
SJ <shivanijain.shivanij...@gmail.com>
|
What is a good use case for accepting 1 of N transfer syntaxes from an A-ASSOCIATE-RQ
An SCU can request more then 1 transfer syntax in a presentation
context. For example: JPEG Lossless, Explicit Little Endian and
Implicit Little Edian. The SCP must select one of the proposed
transfer syntaxes or reject the requested abstract syntax.
A question is which algorithm should the SCP use to select which
transfer syntax to accept?
- A) The SCP selects the first transfer syntax that it can support.
The SCU can anticipate on this behavior and sort the requested
transfer syntaxes in a certain way.
- B) The acceptor has its own preference. It selects for example a
lossless
|
3/21/2010 4:16:59 PM
|
0
|
Victor Derks <vbade...@gmail.com>
|
How to send patient info to medical device
Hi,
I need to write a simple application in c# which will communicate with
medical device and send to this device information about patient
(name,age,id,etc). How to do this with DICOM ?
I know that application 'Rislink Server' can do this, but I need to
write my own app.
This simple device is a Cawo Digidcam, used to digitally print patient
data on X-Rays.
I will be very grateful for help or any suggestions.
|
3/20/2010 9:26:55 PM
|
0
|
"Mike P." <palla...@gmail.com>
|
Possible to determine T1 vs. T2 from DICOM tags?
Greetings,
I am trying to extract as much data as possible from both DICOM and
HL7 (study description, series description, etc.) for all images
stored on our PACS, in order to further manipulate the data.
I understand that the response times vary by tissue, but is there a
"foolproof" way to distinguish T1 from T2 by dicom data alone? In
other words, is there a standard calculation that can be made on:
(0018,0080) RepetitionTime
(0018,0081) EchoTime
In order to flag a series correctly, or do I need to rely on the
series description .. (and if so, is the series description a re
|
3/19/2010 3:18:40 PM
|
0
|
hangingProtocol Research <hangingprotocolresea...@gmail.com>
|
Help clarify DICOM orientation and position
Hello all,
Please excuse me for bumping a 15-year old post, but I just came
accross this post from cperz, which looks like it was never
answered .. but is similar to my own questions .. so i'm 'bumping' it,
in the hopes that there might be some clarification on the matter.
FYI: original post is here: http://groups.google.com/group/comp.protocols.d=
icom/msg/9ba7210c13048130?dmode=3Dsource
------------------------
I have interpreted the DICOM orientation (0020,0037) and image
position
(0020,0032) fields and wonder if anyone can verify or correct my
results. The following tab
|
3/19/2010 2:20:05 PM
|
0
|
hangingProtocol Research <hangingprotocolresea...@gmail.com>
|
Localizer Lines
Hi, i m Cantek =C3=87ET=C4=B0N.
I want to ask specially to David CLUNIE.
i m trying to add referance lines on my project but i couldnt get it.
i read your writings but there are some points that i dont understand.
I want to draw the lines on referance image so i need at least 2
points, but after your calculations i m getting only one and i think i
m getting it wrong !
here is your calculations;
dst_nrm_dircos_x =3D dst_row_dircos_y * dst_col_dircos_z -
dst_row_dircos_z * dst_col_dircos_y;
dst_nrm_dircos_y =3D dst_row_dircos_z * dst_col_dircos_x -
dst_row_dircos_x * dst_col_d
|
3/19/2010 10:15:45 AM
|
0
|
circass <circ...@gmail.com>
|
exception establishing association- AETitle scuname is already declared.
Hi,
I am trying to create an association from a SCU to SCP. I am getting
an exception
AETitle scuname is already declared.
Can someone tell me how to resolve this? I haven't already established
association with this SCU (AETitle: scuname). Any ideas?
Thanks in advance.
|
3/19/2010 3:15:44 AM
|
0
|
dicom_user <shakthi.1...@gmail.com>
|
Cmove Dicom code
Hi Everyone,
I am newbie to dicom. I am trying to create a SCU that will establish
an association with SCP and send a cmove request. I am pasting the
method code for doing the cmove request in java. Can someone who is
familiar or has done this please verify if my cmove code is correct? I
would like to get some feedback on if I am missing any steps. Any
feedback is appreciated.
protected MCdimseMessage performCMove(final MCassociation a, final
MCdataSet data) {
MCdimseMessage rq = null;
if (data != null) {
rq = doSendServiceRequest(a, data);
rq.setServiceCommand(MODALI
|
3/19/2010 3:05:37 AM
|
0
|
dicom_user <shakthi.1...@gmail.com>
|
New DICOM image viewer for Windows
Hi,
RadiAnt Viewer is a freeware DICOM image viewer for Windows systems,
32 and 64 bit versions are available for download. It is in early
stage of development, currently allowing image series browsing,
zooming, moving, window adjustment, length and density measurements.
It was designed to be lightweight and efficient. Comments greatly
appreciated.
http://www.RadiAntViewer.com
|
3/1/2010 7:39:31 PM
|
0
|
Maciej Frankiewicz <radiantvie...@gmail.com>
|
Worklist from RIS question
I have a question regarding scheduled procedures (we use RIS generated
worklists)
Patient "A" and "B" are scheduled for an exam and patient "A" does not
show up for the exam, and reschedules for 5 days in the future, RIS
uses the same Study UID (does not cancel the order, meerly reschedules
the order). In the meantime the technologist has used Patient "A"
scheduled exam by mistake on patient "B", tech realizes mistake and
moves images to patient "B", technologist merges the correct order in
PACS that it has for patient "B" (Study UID is one that was scheduled
for patient "A") Patie
|
3/1/2010 1:34:18 PM
|
0
|
bd6reds <bd6r...@gmail.com>
|
Philips Dicom Designers Cookbook
I am looking for this Philips Dicom Designers cookbook, but can't find
it anywhere. There is another Philips document that was on their
connectivity web page that I am looking for. Anyone have a copy of
these that they can email to me?
Thanks
Martin
|
2/28/2010 11:24:33 PM
|
0
|
"Martin Riddle" <martin_...@verizon.net>
|
Call for Papers - DICOM 2008 International Conference & Seminar
DICOM 2010 INTERNATIONAL CONFERENCE & SEMINAR
Call for Papers
9 - 11 October 2010
Rio de Janeiro, Brazil
The DICOM 2010 International Conference and Seminar -- held in
conjunction with the 2010 Brazilian Congress of Radiology -- will
discuss the current, anticipated, and future features of the DICOM
Standard and its role in health care.
- Invited lectures will discuss current features of the DICOM standard
and how they interplay with each other and the rest of the medical
world.
- The conference will include oral and poster presentations on new and
upcoming DICOM features a
|
2/27/2010 7:23:16 PM
|
0
|
Chuck Kahn <cek...@gmail.com>
|
JPEG Lossless YBR_FULL
Is it legal to encode YBR_FULL images in JPEG Lossless?
My interpretation of the standard says it isn't, specifically:
PS 3.3 C.8.5.6.1.16 says the US Planar Configuration
YBR_FULL Planar 1 - colour by plane
but compressed formats excluding RLE have to be interleaved.
As well, the photometric interpretation needs to be consistent with
the compressed data format requirements (according to PS 3.5), but I
was unable to figure out what the JPEG Lossless header can specify in
terms of photometric interpretation. If it only specifies grayscale
or rgb, then YBR_FULL would be disallowed,
|
2/26/2010 3:20:14 PM
|
0
|
Bill Wallace <wayfarer3...@gmail.com>
|
DICOM image transformations
Goodmorning, I need to ask if the transformations (change grayscale,
rotation, contrast,..., or adding ROI or any other indication in DICOM
file) is normally performed by viewers or any other program that
manages DICOM, are normally write as tags or if image data are
changed?
Are there tags that can describe all the normal transformations on the
image? If yes, viewers are able to display the correct transformations
reading these tags?
Thanks for any answer..
|
2/25/2010 11:40:19 AM
|
0
|
signo88 <mirko.signor...@nsbusiness.it>
|
Presentation / Transfering the image over network
Hi
DICOM Images are in the PACS for each studies.
The Images may be having the annotations / Presentation states .
The radiologist will review the cases through a DICOM Viewer.
Transfering the image as in the DICOM format is difficult as it may be
having comparitievely large file size .
We know that the image can be send in the non dicom format with the
wado.
While sending in the jpeg format is it possible to have the
presentation details / annotation on the image
And is it possible to convert the jpeg to the DCM Image at the client
side.
Thanks in advance
Regards
R.Josep
|
2/24/2010 8:23:46 AM
|
0
|
"Robin.." <robinjoseph...@gmail.com>
|
Transfering DICOM Images over network
Hi
I am now in the initial phase of a teleradiology project.
The radiologist have to review the case from any location in the
world.
The DICOM image may be in the PACS server and the radiologist have to
view it through the DICOM Viewer.
As the DICOM Images may be having larger file size , it will be little
difficult to transfer over the network.
We are thinking about the secured VPN for this .
Is it a better one fro the transfer of the DICOM Images?
Can any one suggest a better option for the DICOM Image transfer over
network.
Thanks in advance
R.Joseph
|
2/24/2010 8:12:11 AM
|
0
|
"Robin.." <robinjoseph...@gmail.com>
|
DICOM files "OT" modality
Hi
I just start to learn DICOM and need to export a few files in DICOM
format.
I create a wrapper DICOM files with the "OT" modality and attached a
file in one of each private field in each DICOM files, what other
attributes must I declare (add to these files) to send them to a PACKS
server ?
Appreciate any help.
Ran
|
2/23/2010 6:49:56 PM
|
0
|
"R.G" <ran...@gmail.com>
|
How to start with gdcm
Hi,
I am a complete newbie to gdcm. I want to use gdcm to extract some
header data and a single image from dicom files.
I thought it would be a good start to try out the HelloWorld example
from the gdcm documentation. My problem is that I don't know what
libraries or Dlls to include from the gdcm installation. I installed the
precompiled version for windows. Can anyone point me to the right
direction, please?
thanks in advance
Stefan
|
2/23/2010 12:56:33 PM
|
0
|
=?ISO-8859-15?Q?Stefan_Gro=DF?= <stefan.gr...@stud.tu-ilmenau.de>
|
distinguish between real CT scans(beam ON) and non-real CT scans(beam OFF)
Hi,
I am trying to recursively read all dicom images from a CT study with
several series. How do I figure out which images are real/primary
scans(where the x-ray beam was ON) and which ones are not( such as
recons, etc.)? I have been looking at the Image Type Tag(0008,0008)
and series description tag (0008,103E), but is there a more systematic
and foolproof way of figuring this out?
I would appreciate if anyone can help me figure this out.
Thanks,
Ashish
|
2/19/2010 6:49:08 PM
|
0
|
Ashish Singh <asinghmli...@gmail.com>
|
Looking for exhaustive list of "standard" hanging protocols, by modality
Hello,
I work for a developer of radiological PACS and diagnostic viewing
products.
Our viewing software already has a decent method for customizing
hanging protocols, but we're looking to make it a lot better.
As part of our research, I'm hoping to find a list that details common
(or popular) layout, or reading styles based on modality, body part,
etc.
i know this will be a massive list, and will vary greatly from
radiologist to radiologist, and this is precisely part of our
challenge. we have begun to interview our clients, but every scenario
is so unique, we want to be su
|
2/19/2010 3:01:45 PM
|
0
|
renderedlife <renderedl...@gmail.com>
|
C-Store, multiple transfer syntaxes
Hi guys,
Maybe somebody has an answer to my question. The SCU requests an
association with 2 presentation contexts, both of the same abstract
syntax but with difference transfersyntaxes. F.e. CT with lossless
JPEG and CT with baseline JPEG. The SCP accepts both and now the SCU
issues a C-Store with 2 files, both CT but 1 file is lossless
compressed and one is baseline compressed.
How does the SCP know which one is lossless and which is lossy? Is it
up to the SCP to detect the compression used in the pixel data?
Thanks,
Rutger
|
2/19/2010 1:36:27 PM
|
0
|
Rutger den Boer <rutgerdenb...@gmail.com>
|
Unique identification of patient
Hi,
I am a newbie in RIS. When developing a RIS, we
need to give an unique identification for each patient. I would like
to know how this can be achieved. Can anybody help me. I have read in
few articles that a MRN number is generated for each patient during
the registration time. But what will happen if patients on different
hospital have the same identification number? Hope you understood what
i am asking.
Thanks and regards,
Sreelash
|
2/17/2010 5:18:22 AM
|
0
|
sreelash s <sreelash.br...@gmail.com>
|
WADO image/jpeg
According to PS 3.18-2009 "Web Access to DICOM Persistent Objects
(WADO)" images of MIME type "image/jpeg" shall be encoded "using the
JPEG baseline lossy 8 bit Huffman encoded non-hierarchical non-
sequential process ISO/IEC 10918".
As far as I can tell the baseline process is "sequential" which would
conflict with "non-sequential".
Non-hierarchical, non-sequential, lossy would imply progressive
coding; is this what is meant?
Also the standard does not seem to say which file format to choose.
RFC2046 defines "image/jpeg" to be JFIF ("An initial subtype is "jpeg"
for the JPEG f
|
2/16/2010 10:54:25 AM
|
0
|
ngladitz <glad...@sci-vis.de>
|
C-FIND RQ without matching keys
What's the right behavior of a DICOM C-FIND SCP (of the Composite
Object Query/Retrieve Service Class, the Modality or the General-
Purpose Worklist Management SOP Class , the Substance Administration
Query Service Class, or the Hanging Protocol Query/Retrieve Service
Class), receiving a C-FIND RQ with an identifier with only contains
Key attributes without values?
The question is raised by the sentence
"If there are no matching keys, then there are no matches, return a
response with a status equal to Success and with no Identifier."
in the specification for such C-FIND SCPs. So
|
2/15/2010 11:54:07 AM
|
0
|
gunter zeilinger <gunte...@gmail.com>
|
Edge enhanced images
Hi,
If an image is 'edge-enhanced', how should the image be saved?
Is it possible to just save the Kernel & Filter in any of the standard
DICOM tags (image or PR or any other object)?
Thanks,
SJ
|
2/14/2010 3:48:31 PM
|
0
|
SJ <shivanijain.shivanij...@gmail.com>
|
DICOM Printer
I am looking for a DICOM printer with support for almost all SOP
classes.
Can anyone please suggest
Thanks and Regards,
Sanjay
|
2/11/2010 4:42:52 AM
|
0
|
Sanjay <sanjay....@gmail.com>
|
sample image for modality,voi and presentation lut
Hi ,
Could you please send me a DICOM image or a link where i can download
a DICOM image which has MODALITY LUT, VOI LUT and PRESENTATION LUT all
together present in a single image.
jagadish
|
2/9/2010 11:03:02 AM
|
0
|
Jagadish <jagadis...@gmail.com>
|
Interpreting DICOM Pixel Data Type.
How do you go about interpreting the VR of Pixel Data.
Particularly in an implicit environment.
There seem to be a number of ways to interpret these keys
Bits Per Pixel
Samples Per Pixel
High Bit
Bits Allocated
Bits Stored
Pixel Representation
Photometric Interpretation.
Where in the standard can I find a discussion of this. I looked
through part 3 and found the elements.. but...
Should I have been looking at the SOP Class definition? Would it
differ between SOPs?
Is there some way to tell if a 32 bit value is a float or a long
integer.... Stuff like that....
|
2/8/2010 7:51:29 PM
|
0
|
SpreadTooThin <bjobrie...@gmail.com>
|
MODALITY LUT,VOI LUT,PRESENTATION LUT
Hi guys,
I got a doubt in DICOM tags.Is it possible that an image can have
MODALITY LUT, VOI LUT & PRESENTATION LUT all together in a single
image.Why i am asking is because Tag for LUT data is (0028,3006) which
is same for all the 3 LUT's(Modality,Voi,Presenatation LUT).
For example lets take that we have an image which has both Modality
LUT and VOI LUT, in such a case (0028,3006) tag will have which
data,Modality LUT Data or VOI LUT Data?.
Jagadish
|
2/8/2010 11:38:46 AM
|
0
|
Jagadish <jagadis...@gmail.com>
|
DCM4CHEE suppports Presentation State SOP ?
I would like to send the image and the softcopy Presentation state to
the PACS
I am planning to use the dcm4chee as the image archieve.
Before that i would like to confirm that :
"Does dcm4chee support the Presentations state SOP classes ?"
Thanks in advance
Regards
Robin
|
2/8/2010 6:03:30 AM
|
0
|
"Robin.." <robinjoseph...@gmail.com>
|
DICOM Analyer
Here a good tool to analyze your dicom connections (DICOM store, send,
verification, ...).
You can find it here:
http://www.dobco.eu/software.php
|
2/7/2010 3:21:43 PM
|
0
|
Wesly Wes <dobco...@gmail.com>
|
Significant change in pixelmed and dicom3tools: keywords updated per CP 850/2009
CP 850:
"ftp://medical.nema.org/medical/dicom/final/cp850_ft.pdf"
finally added rules for generating keywords for all data elements, and
this was
instantiated with defined keywords in the 2009 edition of PS 3.6.
Accordingly, I have updated both dicom3tools and the pixelmed toolkit
to use the
new keywords in place of those used previously.
This will affect both scripts that use the command line tools with
keywords, as
well as code that uses the keywords to look up tags in the dictionary.
In the pixelmed toolkit, the database column names will also change in
some
cases.
|
2/7/2010 1:48:34 PM
|
0
|
David Clunie <dclu...@dclunie.com>
|
DICOM Status Page Updates 2010/02/06 after 2006/01 WG 6 Meeting
As of Sat Feb 6 14:09:50 EST 2010 the DICOM status page on my web site has been
updated with the following changes after the Jan 2010 WG 6 meeting. See:
http://www.dclunie.com/dicom-status/status.html
Note that the status and document references for supplements
and correction proposals reflects the documents that have
been made available by their editors during or after the
most recent meeting; hence the status may be out-of-date relative to
the official status when the most recent documents have
not yet been made available.
The following supplements that were adopted a
|
2/6/2010 7:11:12 PM
|
0
|
David Clunie <dclu...@dclunie.com>
|
Best Image Processing Open Source API in JAVA
I would like to create DICOM Image viewer and also want to implement
DICOM services like C-FIND,C-MOVE etc to communicate with the PACS
I would like to use the API for both the viewer and DICOM services.
Could any one please tell me which is the best open source API in
JAVA
Thanks in advance
Robin Joseph..
|
2/5/2010 6:54:39 AM
|
0
|
"Robin.." <robinjoseph...@gmail.com>
|
Saving Images back to PACS/ Altering the Image
The radiologist is reviewing / interpreting the DICOM image which is
retrieved from the PACS
He have added annotations in the images and is it good to save the
image back to the PACS .
I have read that
The base line in DICOM is once an image has been created, it should
NEVER be changed.
So is it against the DICOM standards to alter the image which is once
created .
Also is there any provision to save the DCM image with annotations and
markings..
Thanks in advance..
Robin Joseph..
|
2/5/2010 4:44:41 AM
|
0
|
"Robin.." <robinjoseph...@gmail.com>
|
rtstruct series UID
Hi,
I have a question regarding the series UID of dicom rtstructs.
It seems to me that the series UID of an rtstruct should match the
series UID from which the segmentation was made. For instance, if I
segment off of a CT image, then the rtstruct that's produced should
have the same series UID as that CT Image, but a different instance
UID (to indicate that it's a different bit of information in the same
series).
I mention this because I'm trying to create rtstructs using this code
here:
http://www.insight-journal.org/browse/publication/316
However, they are creating their
|
2/4/2010 5:13:20 PM
|
0
|
mmr <mmro...@gmail.com>
|
velocity encoded mri
Hi,
I have 2D and 3D mri fils. I want to see the velocity values. But I
have some problems, such as
for the 3D dataset the image details are, 'velocity direction 0/0/0
minimum velocity 0'. So I think I need to make some parameter settings
for image acquisition. But I don't know how to.
Could you please help me??
Best regards.
|
2/4/2010 10:18:27 AM
|
0
|
Gizem inci <gizemi...@gmail.com>
|